avatar: Biomed 2
Biomed 2
1 posts
Replied to Biomed 2's post on November 25th, 2014

I can'nt move these knobs to adjust the stage they feel to be stuck does this ment the knobs are bad can new knobs be install

avatar: JAC47
JAC47
27 posts
Replied to Biomed 2's post on December 1st, 2014

By clamp knob do you mean the tension adjustment ring that is located on one sde between coarse focus and microscope body. It works by increasing friction. Hold the coarse focus knob and try to turn the tension ring. Some scopes have locking set screws in tension ring so you may need to loosen them slightly with correct sized metric driver.

avatar: Duluth6
Duluth6
14 posts
Replied to Biomed 2's post on August 26th, 2015

If you can't move the stage up to the focal point at 10x.The refocus stop on the oposite side of the tension adjustment side may be locked in place or the friction plates may be sticking. The tension adjustment on the other side maybe to tight. Please do not turn the coarse focus knobs aginst each other- doing so will break things. The e400 is out of production- Nikon still has a few parts.

avatar: Duluth6
Duluth6
14 posts
Replied to Biomed 2's post on December 2nd, 2016

The refocus assembly needs to be rebuilt- no parts ~$120 provided that is the only problem.