I am trying to execute an autotune using the syringe pump. However, there is no mist spray from the needle. I have checked for blockage in the line, and I can confirm there is none. Can anyone offer advice on what I could do. I allowed the equipment to pulse for >1hr to allow it to thermally stabilize, and I still dont see all ions of interest.

avatar: Vanessa C.
Vanessa C.
2 posts
Replied to eddy's post on December 5th, 2017

A few things I've tried in the past:

1. use new calibration fluid

2. use the other ESI port needle

3. check the gas compresser