No Air Pressure (Interlock ERROR)
PerkinElmer AA 400. I am getting this interlock error "No Air Pressure" but the air supply is active. Where can I look for the pressure sensor. It seems that this could be the problem. Is it on the flow block where the gasses are connected?
How do I fix the "drain is not connected" error, when the drain is physically attached?
There can be three problems:
1. Check if drain interlock cable connector is connected
2. Your drain vessel could be filled with solvent. Empty the drain vessel but you should leave a little bit of water in vessel
3. Disconnect the drain tubing and shake your drain vessel
Did everything, plus added some water through the drain, and it worked! Appreciate your help.
Try the following.
1. Check that the air pressure is sufficeint.
2. Make sure the nebulizer gas hose is connected and not loose.
3. Bleed gas lines
If the interlock error remains, check the drain system.
1. If the drain bottle is full, empty it.
2. The loop may need water. Disconnect the drain hose from the end cap. Pour 500 ml DI water down the drain hose. Replace the hose.
If you still have problems e-mail AAtechsupport@perkinelemr.com