i have a Karl fisher of mettler toledo V20 problem is that when we drain the sample it automatically fill the methanol.Please rectify this issue.

avatar: jburkman11
jburkman11
155 posts
Replied to Hashu's post on October 11th, 2017

If you are using the drain function and the solvent manager is also filling, then it is not operating properly. There may be a faulty or damaged valve. If the valve is damaged, you would need to replace the solvent manager.

avatar: Hashu
Hashu
2 posts
Replied to jburkman11's post on October 12th, 2017

We replaced the solvent manager with new one but still problem is same.

avatar: jburkman11
jburkman11
155 posts
Replied to Hashu's post on October 12th, 2017

Please contact Mettler Toledo Lab Tech Support at 800-638-8537, X103344