Hi everyone,

just replaced the board and got the error message IMBAL when attempting to run a program.

With the old board there was no IMBAL error.

I wonder if one needs to "program" the board before installing?

thank you,

Miguel

avatar: medphysteo
medphysteo
258 posts
Replied to miguelR's post on December 7th, 2017

Hi Miguel,

You do have to recalibrate the imbalance detection for the new board after installing. See

http://www.labwrench.com/?community.posts/threadNo/163265/subject/Imbalance-error-after-changing-main-board/

Jeff

avatar: miguelR
miguelR
19 posts
Replied to medphysteo's post on December 7th, 2017

Hi Jeff, 

thanks a lot for your reply. I try to follow the forum as much as possible but I guess I missed that topic :/  

When you instal such board, do you check or calibrate any other thing besides the imbalance? 

Thank you so much, 

Miguel 

 

 

avatar: medphysteo
medphysteo
258 posts
Replied to miguelR's post on December 8th, 2017

Hi Miguel,

You're welcome.

The imbalance detection is the only thing you have to reprogram for a new board.

But, a new board is not the only time you have to reprogram the imbalance detection. You also have to do it when you replace the motor or motor mounts, the rotor, and of course, the imbalance detection board.

Jeff