We are seeing an error this morning stating "Error MD1101. Flame has gone out - Check the fuel and oxidant pressures."

There are no leaks on the gas line.

Flame is igniting but shutting off and showing that error when we proceed to calibration step.

Any ideas of what could be going on here...

avatar: Mathias
Mathias
1 posts
Replied to Kelesha's post on February 13th, 2018

I have the same problem. I've found out so far that it shuts off the fuel gas supply while the flame is still on. I think he doesn't realize the flame is on. Behind the flame there is a small window with a small detector which should register that the flame is on. A possible mistake would be that it is defective.
If you hold down the ignition button, even though the flame is well lit, it will continue to burn without any faults until the ignition is automatically stopped for reasons of time.
If there are any new findings, I would also be very grateful for this.
Best regards
Mathias