Replied to 20120620115648077's post on June 28th, 2012

Dear Manager,

What would be inflow speed (fpm) for #1186?  If inflow is too high, can we just the damper to lower the flow?  Thanks. 

avatar: Certification
Certification
11 posts
Replied to 20120620115648077's post on September 11th, 2012

Inflow should be 85-95.   Hoods need to meet an operating range for inflow and downflow.  Adjusting just the speed controller may cause other parameters to be out of specification.  Therefore it would probably be best to allow a trained certifier make adjustments upon certification.