The pump is on abut is not pumping liquid/sample, does anyone have any ideas on how to remedy this?  What could be the cause?  Thanks, Mohamed

Replied to MHOMAR's post on July 11th, 2012

Check your pump tubing to make sure it's not flattened. Also check the tension on the pump cassette to be sure it's not too tight or too loose.

Replied to MHOMAR's post on July 11th, 2012

Thanks 20110705111537533   that is just what we would recommend for Mohamed.  Mohamed if that doesn't fix your problem, please feel free to contact our customer support directly at 1-800-533-6267. Thanks Teledyne Leeman Labs