avatar: Rugi_28
Rugi_28
2 posts
Replied to Rugi_28's post on August 27th, 2018

I am using the PU 1580 HPLC pump in my set-up.  Every time I turn it on the pressure sensor always displyas :"error" which consequently I cannot use it at that instant.

How can I fixed this error display?

 

Thanks

avatar: JRB
JRB
11 posts
Replied to Rugi_28's post on August 27th, 2018

the error is caused by residual pressure at power on or the sensor being out of calibration. Open the drain valve and restart the instrument. If the error persists, it is likely that the pressure sensor is out of calibration.

There is an adjustment screw on the rear panel to zero the 'true' reading. If the value is out of calibration the error message will be displayed.  Press clear and the instrument will continue initializing. Press shift and then 6 and you can see the true value, carefully adjust the rear panel using a small screwdriver until the value reaches zero. Power off and on.

If the error persists it is likely the pressure sensor has failed and needs to be replaced.

avatar: Rugi_28
Rugi_28
2 posts
Replied to Rugi_28's post on August 27th, 2018
Hello JRB. May I ask where is the locatio of the the drain valve? Thank.