hi >

the problom I have contamination effacting my result , in BIOTEK ELx 50 microwell wosher.

how do I solve this problom?

avatar: PatrickD
PatrickD
5 posts
Replied to wsiba's post on September 14th, 2018

Replace all tubings.

Decomantamine tip manifold. Basically clean the entire liquid path.

avatar: wsiba
wsiba
3 posts
Replied to PatrickD's post on September 14th, 2018

do I have to use spacific chemical Solution to clean the path ?

avatar: PatrickD
PatrickD
5 posts
Replied to wsiba's post on September 14th, 2018

Try 70% ethanol and purge with sterille water

avatar: wsiba
wsiba
3 posts
Replied to PatrickD's post on September 14th, 2018

thank U for UR information . hop this work's