Replied to Jkirk's post on August 16th, 2019
Thermo Scientific Sorvall™ Legend™ Micro  21R Microcentrifuge error 24 at power up. Are there any fixes besides replacing the control board and display?

Does anyone have a copy of the service manual?

avatar: RK21702
RK21702
4 posts
Replied to Jkirk's post on July 30th, 2020

Turn centrifuge off, manually open lid by pushing a heavy paper clip through small hole that is in front of the lid, look down inside the latch so you can see the lid switch, switch centrifuge on..push lid switch down with a small screwdriver then release...then down again then release. Remove 4 screws to lift off top case then remove latch assembly and check micro switches.