EPPENDORF 5804R Centrifuge, every time we want to spin something down ERROR 1. 

 

PLEASE HELP

avatar: WarwickJon
WarwickJon
88 posts
Replied to BaeLab103's post on February 5th, 2020

Error 1 reads an error in speed measuring system.  The manual suggests trying a different rotor.

 

https://www.nist.gov/system/files/documents/ncnr/Eppendorf_5810R_Centrifuge_Manual.pdf

avatar: BaeLab103
BaeLab103
4 posts
Replied to WarwickJon's post on February 5th, 2020

Yes, it doesn't happen with another rotor. But I need to use this rotor, so what do I do?

avatar: WarwickJon
WarwickJon
88 posts
Replied to BaeLab103's post on February 6th, 2020

Is the rotor you are trying to use approved for use in this centrifuge?  If so, the unit isnt recognising it.

avatar: BaeLab103
BaeLab103
4 posts
Replied to WarwickJon's post on February 6th, 2020

Yes, it is approved for use in this centrifuge. 

What can I do to make it recognize it / properly measure the speed?

avatar: WarwickJon
WarwickJon
88 posts
Replied to BaeLab103's post on February 7th, 2020

If the rotor is a newer type maybe the firmware needs an update?  May be worth checking with Eppendorf.

avatar: BaeLab103
BaeLab103
4 posts
Replied to WarwickJon's post on February 7th, 2020

Thanks, will do.