Replied to Deleted User's post on June 9th, 2011
Error 6 has numerous causes, this unit was working until moved, now it's all error 6. Temp control works fine. Does anybody know if the service manual is worth the $$? I don't want to replace the freq conv bd if I can see a work around. I have several other svc manuals for eppendorf, but the tech support folks are so unhelpful, that I may just recommend replacing unit with another more servicable brand.
avatar: Sime
Sime
1 posts
Replied to Deleted User's post on June 15th, 2011
Hi, the service manual states that there are 4 probable causes for Error 6:
- convertor is overloaded; allow unit to cool and retry after 5 minutes
- the brake is defective; no remedy given in the manual for this!
- the voltage supply is too low; check voltage supply to unit
- the rotor has come loose; retighten rotor

If the unit has just been moved, first thing I'd try is removing then refitting the rotor. If this doesn't work I'd check the voltage. If its been left switched on for a long time it might not have appreciated being switched off while it was being moved, so it might have blown something on the voltage regulator (where the power cord goes into the back of the unit).

I hope you manage to sort it out - my 9 year old 5810R died last month due to the rotor sensor failing. Eppendorf claimed they only keep spare parts on equipment for 7 years and since mine was 9 years old, no spare parts were available. Unfortunately, I had to scrap the centrifuge. Hope you have more luck with yours!
All the best
avatar: Deleted User
Deleted User
posts
Replied to Sime's post on June 15th, 2011
Thank You- I did check those items, and actually had another unit with the same error code, which wasn't moved, and also had the freq conv board replaced less than a year ago. I wish eppendorf wasn't so 'black-box'. I have tech info on almost all other brands, but they're all about selling machines, not parts. It is unrealistic to offer either ship-in or exorbitant service only. Our clients are finally understanding that they don't want centrifuges that you can't repair economically. I can't wait to see how New Brunswick changes.
avatar: NITS
NITS
9 posts
Replied to Deleted User's post on October 2nd, 2012

Were you able to fix the error 6 on your machine? Please let me know. Also, do you have the service manual?

Many thanks.

avatar: Reno
Reno
3 posts
Replied to NITS post on October 16th, 2013

Hallo, I have the same problem - error 6. Does anybody have service manual to send me? Thanks!! Reno , My adress E-mail: reno.sime@gmail.com

avatar: LabWrenchSiteAdmin
LabWrenchSiteAdmin
991 posts
Replied to Reno's post on October 16th, 2013

Hello Reno

LabWrench has a partial User Manual for the Eppendorf 5804 / 5804R / 5810 / 5810R Centrifuges posted on the DOCUMENTS tab of the product page.

Below is a direct link to the User Manual. 

http://photos.labwrench.com/equipmentManuals/2402-202.pdf

The error codes are noted on page 41.

Please feel free to contact me if you have any questions.

Respectfully
Ronnie Stanley
 

avatar: Reno
Reno
3 posts
Replied to LabWrenchSiteAdmin's post on October 16th, 2013

Hallo Ronnie,

thank you very much for User manual!

Do you have service manuel too?

Thanks and regards

Reno

avatar: LabWrenchSiteAdmin
LabWrenchSiteAdmin
991 posts
Replied to Reno's post on October 16th, 2013

HI Reno

No we do not have the Service Manual here.  There are many publications on the Eppendorf Website:

http://www.eppendorf.com

Click on the following links

1. Support and ep Services

2. Literature Library

3. Manuals

There you will find a list of products to choose from.

Respectfully
Ronnie
 

avatar: Augusto
Augusto
2 posts
Replied to LabWrenchSiteAdmin's post on November 11th, 2013

Hello

 

Sorry for my english .

Usualy the problem ( in the case of error 6  - over load ) it's in a IGBT conected betwen the bridge rectifier and the brake resistor located in the inverter module.

This module it's the module conected to the motor.

 

sorry my english but today i've repaired another centrifuge with the same problem.

 

Please don,t hesitate to contact me.

 

All the best

Augusto hospassis@gmail.com

avatar: Reno
Reno
3 posts
Replied to Augusto's post on February 7th, 2014

Hi,

I replaced two biggest electrolytic capacitor on  frequency convertor board - it solved problem. (there was error 6)

  regards   Reno Zagreb     

avatar: sarwan
sarwan
1 posts
Replied to Reno's post on July 6th, 2014

Hi Guys

I have a fault with this centrifuge with lid lock issue. When trying to close the top lid, the motor tries to latch the lid but stops half way. after few tries it does close, but fails to open the lid when button is pressed. I have replaced the lid lock motor and the lid microswitches but fault remains. I have monitored the motor voltage. there motor works without load, but voltage cuts off as soon as load is applied to it.

any suggessions on what might be causing the motor voltage to cut off??

also link to service manual would be helpfull.

 

thanks

Sarwan

avatar: medphysteo
medphysteo
230 posts
Replied to sarwan's post on July 7th, 2014

Here are links to the service manual. Sorry I had to break it into 3 parts, due to email server limitations.

Centrifuge 5804/5804R Centrifuge 5810/5810R Refresher (1 of 3)

Centrifuge 5804/5804R Centrifuge 5810/5810R Refresher (2 of 3)

Centrifuge 5804/5804R Centrifuge 5810/5810R Refresher (3 of 3)

avatar: medphysteo
medphysteo
230 posts
Replied to NITS post on July 7th, 2014

Service manual is now posted in 3 parts here:

http://www.labwrench.com/?equipment.view/equipmentNo/2402/EPPENDORF/5810-5810R/

avatar: sarjpatel
sarjpatel
1 posts
Replied to Augusto's post on August 1st, 2014

Augusto,

Do you have a diagram to illustrate how to make this repair.  I have a 5810R that is throw me an error 6.  The centrifuge continue to cool down and everthing else appears normal unitl i start a run and it indicates an error 6??? could you help me out?

Sincerely,

Sarj

 

avatar: jay.patel
jay.patel
1 posts
Replied to medphysteo's post on March 23rd, 2015

Quote medphysteo

Service manual is now posted in 3 parts here:

http://www.labwrench.com/?equipment.view/equipmentNo/2402/EPPENDORF/5810-5810R/

 

 

You are awesome.  Was looking for the service manual everywhere for my centrifgue 5804.

avatar: jjj
jjj
7 posts
Replied to medphysteo's post on June 15th, 2015

Does anyone know where I can get a schematic of the Control card for the 5810R?

avatar: wsmith
wsmith
1 posts
Replied to Reno's post on July 8th, 2015

Hello 

Did you ever manage to get a service manual for the 5810R?. and if you did would you be able to send it to me?

 

Thanks

 

avatar: NITS
NITS
9 posts
Replied to wsmith's post on July 8th, 2015

Check the link above, there is a service manual available in 3 parts.

avatar: jjj
jjj
7 posts
Replied to wsmith's post on July 8th, 2015

There are two different versions of the 5810, an older one and a newer one. The manuals from the link given above are for the newer model. I have the older machine and someone sent me the older manual. All of these manuals are scanned copies so the schematic pages are almost impossible to read. Eppendorf sent me the original PDFs for some of these schematics. I forwarded them to the Admin here but I haven't seen them post a link for them yet.

 

My machine gives an error 6. I was able to find that a trace on the main board had burned open. This trace supplies power to the frequency convertor card. Examination of that card shows a very slight deforming of the two large capacitors on it. I can't readily see how to seperate this card from the metal base it is attached to. Reno in a post above says he replaced two capacitors to fix his machine.

 

Hey RENO! If you're listening can you tell me how you did that?

avatar: NITS
NITS
9 posts
Replied to jjj's post on July 8th, 2015

Just to give you a trick, i had to replace a part and the metal base was driving me nuts. I drilled two holes right behind the connectors and resoldered the part. Worked fine and i dont think there is any other way.

avatar: MSUStevo
MSUStevo
2 posts
Replied to jjj's post on August 13th, 2015

jjj, would you mind posting the service manual for the older 5810R? We are currently servicing a unit showing Error 6, and would find the manual very helpful!

avatar: Geords
Geords
12 posts
Replied to MSUStevo's post on August 18th, 2015

jjjj, can please post the service manual for the older version? It would also be helpful to me.

Thsanks in advance 

Geords

avatar: jjj
jjj
7 posts
Replied to Geords post on August 18th, 2015

I sent the manual to Admin. If you post an email I will try to send it to you directly.

 

avatar: MSUStevo
MSUStevo
2 posts
Replied to jjj's post on August 25th, 2015

Hey jjj, you can send it to

brookss6 at msu dot edu

Thanks, Steve

avatar: Alexan
Alexan
1 posts
Replied to Reno's post on December 5th, 2015

Hi 

My centrifuge 5804 R  showing error 2. The only way to run the centrifuge is to run instantly after closing the lid. otherwise, after 2-3 seconds of closing the door the error 2 comes and not allow to start centriguge.

Is there a way to reset error 2?

Thanks

Alex

avatar: service3683
service3683
19 posts
Replied to Alexan's post on December 5th, 2015

Dear Alexan. Your error is related to imbalance sensor. Check the brown-yellow-green-white ribbon cable connections at the board. You can take it out and in, clean the pins. If not solved do the following: with the centrifuge lid open push both arrow keys, display will change. press up until you see as first digit "6:".  You should have a value of 670 +/- 100 there. If you move the rotor axis this value should change. If the value displayed is out of range, the imbalance sensor should be readjusted.

please check this and post your results in order to help you further.

avatar: mforrest
mforrest
1 posts
Replied to jjj's post on December 16th, 2015

Hi, jjj

Could you send me a link to the old service manual too?

Thanks LF

mforrest at ku dot edu

avatar: marcot
marcot
1 posts
Replied to mforrest's post on January 25th, 2016

Awesome! Thanks a lot for that Service Manual. It just helped us to get a centrifuge running again after an "Error 2" after we had to remove a lot of rust and dirt out of an old centrifuge!

Adjusting the imbalance-sensor still is a huge pain! :)

 

best, Marco

avatar: hxl
hxl
10 posts
Replied to marcot's post on January 25th, 2016

Glad to see the question here because I have suffered this once before. Now I know how to make it. Biotech is always complicated for me. Thanks! 

avatar: jchao
jchao
4 posts
Replied to service3683's post on March 9th, 2016

I have an old 5810R. When power it on, it always ask me to press open lid. Once I open and close the lid, it gives error 2 immediately. I've adjust the imbalance sensor to 667 followed by the service manual, the problem still exist, re-check the balance sensor, still with in the spec 670+/-30.