We have noticed condensation at the bottom of the glass door of the incubator. What could cause this?

avatar: Glenn PICKETT
Glenn PICKETT
63 posts
Replied to ayzx10's post on August 24th, 2012

Hi there,

This could be a poor seal on the outer door, enabling cool air to leak in. Even ambient is cool compared to a chamber at 37°C! Only one degree lower temp than the rest of the chamber is enough to start causing condensation when the relative humidity is around 92-95%. In fact, a 1°C temp drop can induce a 10% humidity drop.

It could also be poor sealing of the inner door, but less likely.

Is the incubator in line with draughts ?

You could try setting the door temperature 0.5°C higher (heated by radiation from the outer door.

Hope this helps.