avatar: BMitchell
BMitchell
4 posts
Replied to BMitchell's post on October 1st, 2012

GFCI receptice trips when the unit reaches 7500RPM. Load reading shows a max of 2 amps. There is no out of balance light or error code. Does anyone have any experience with this problem?

avatar: Ric
Ric
36 posts
Replied to BMitchell's post on October 1st, 2012

The GS15/Allegra21 series is not designed to run on a GFCI outlet, according to Beckman Coulter.  I saw this a lot during my career with them.

Ric

avatar: BMitchell
BMitchell
4 posts
Replied to Ric's post on October 2nd, 2012

The GFCI recepticle was the problem. Thanks very much for the help.

Bruce