Is it possible to replace the sample port? If so, where could I get the part?

avatar: Glenn PICKETT
Glenn PICKETT
63 posts
Replied to ma4534's post on October 17th, 2012

Hi,

You don't say what brand or model you are talking about.

Regards,

avatar: ma4534
ma4534
2 posts
Replied to Glenn PICKETT's post on October 17th, 2012

its a nuaire ir air flow - co2 water jacketed incubator. not sure what the model number is.

avatar: Glenn PICKETT
Glenn PICKETT
63 posts
Replied to ma4534's post on October 22nd, 2012

Hi,

I guess that by sample port, you are talking about a small diameter tube through the wall of the chamber. This is normally used for an independent gas probe , temperature probe or maybe a tubing line to/from a pump

For a water-jacketed model, the sample port is a water-tight integral part of the instrument. I have not heard of one being replaced. Is it corroded?

I suggest that you contact Nuaire. You do not say which country you are calling from. If it is the US, then contact them on 800-328-3352 otherwise (and also possible for US) use the contact form http://www.nuaire.com/contact-us/service-inquiries.php and they will put you in touch with your local dealer.

Hope this helps,