blank/tube thermal control Hybaid express

I have a Hybaid Express used by students. For some reason the thermal Control is not working. I suspect that there is some thermal sensor going Down into the mieral oil that si briken off? Anyone can confirm that there is such a sensor going Down in the mineral oil? Anyone know where to buy a replacement for this part?

Rune.andreassen@hioa.no

avatar placemark

Asked by

Rune Andreassen
Report this Post
avatar placemarkReply by
Bodasafa

Hi, Sthis may be a bit late, but yes you are correct, the probe that is the sample block has a small thermistor inside, they can be easily damaged, ensure the probe is in the oil. I have spares if you require one.

Report this Post
avatar placemarkReply by
Rune Andreassen

Thanks!

Yes, I require one as soon as possible.

Please mail me at rune.andreassen@hioa.no so I can give you further information about my location etc

Kind regards,

Rune Andreassen