I have the alert mYCOALERT with luciferase assayfirefly substrate(kitPromega E1910). What does it mean?

I have test the linearity of the measure :

sample undiluted : 1479481 dil 1/2 :802400 dil 1/4 351486

dil 1/8 156044 1/16 88622. Are they good values. Please could you betweengive me the range of measure of this instrument ( measuring range).If the measure is too high, with what can I dilute the sample?

Yours faithfully

josiane Denis

avatar placemark

Asked by

DenisJ
Report this Post
Bernd HutterReply by
Bernd Hutter

Hi Josiane,

what you did is ideal to judge the linearity and the dynamic range. plotting your data gives an R² of 0.9966 which is perfect. Your samples are all within the range of the instruments

 

bernd