We have a PowePac 300 giving e8

I am trying to find out any information on a PowePac 300. when I set a voltage it will set but when I turn the unit to run I get zero volts out and then e8 pops up in the display. the only information I have found is that e8 means there is a problem with the unit not helpful. if anyone has a schematic or service information or service manual it would really help.

avatar placemark

Asked by

mcneely
Report this Post
avatar placemarkReply by
RK21702

Replace the cooling fan and R10 and R15. If it is a 300CE, both of these resistors are 6.8k 7watt. If its a 300, R10 is 4.7k 5watt and R15 is 2.4k 5watt.