Discussion Category:  Atomic Absorption

Flame :No Air Pressure

When i switch the instument it gives me no air pressure error, i have checked air pressure is ok.

avatar placemark

Asked by

gasimlab
Report this Post
avatar placemarkReply by
marahdi

If you sure that the pressure is ok, may be the problem at pressure switch inside instrument,

Report this Post
avatar placemarkReply by
metro

More than likely it is the drain switch.  pour water in the drain tube so the float switch rises and turns off the interlock.  Hope that helps.