I'm having problems with my 5424 R and was hoping you could help me out here.

The centrifuge runs smoothly as long as it is spinning up or running normally, but as soon as it tries to break, Err 6 appears on the screen. I've already talked to a technician, but they were not too helpful.

Can you tell me how to get into the Err 6 sub menu, so I can see if there's something I can do about this error?

 

greets, ertl

avatar placemark

Asked by

ertl
Report this Post
avatar placemarkReply by
ertl

Me again,

I already fixed it. Just had to open it and solder a wire back on to the power resistor. Now it's perfectly working again.

Report this Post
LabWrenchEditorReply by
LabWrenchEditor

Hi Erti,

Thank you for sharing your solution with us! Always appreciated!

Best Regards,
Kylie


Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com


Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
 
Report this Post
uriminiReply by
urimini

Thanks etrl for sharing this.

 

I am having the same problem. Where should I look to find the power registor?

 

Thanks,

-nj

Report this Post
avatar placemarkReply by
ertl

Hey nj,

 

the power resistor is in the back of the device, it's round, about 7cm long and looks like this: 

 

 

You know that something is wrong with the breaking cycle when the device is working fine while spinning up and holding the speed,  but the moment it tries to slow down, the error appears.

Report this Post
avatar placemarkReply by
jnam

Thanks. This is exactly what happened to my 5424R.

Removing cover was not easy to figure out, though.

Just in case anybody has similar problem: remove lid first and then the big cover.

Cheers.

Report this Post
avatar placemarkReply by
Jonathan Bélanger

HI,

 same issue with mine when starting also !!

can you send me the picture of what the resistor looks like ?

ERTL ?

 

thanks

Report this Post
avatar placemarkReply by
ertl

Quote Jonathan Bélanger

HI,

 same issue with mine when starting also !!

can you send me the picture of what the resistor looks like ?

ERTL ?

 

thanks

 

 

Hey Jonathan,

unfortunately the post is almost 3 years old and the centrifuge has been back at the lab since then.

But the resistor is easily spotable in the back of the device. Do you have a picture of yours so I can mark it?

 

Best,

ertl

Report this Post
avatar placemarkReply by
Jonathan Bélanger

please write an email to me !

Report this Post
avatar placemarkReply by
wlin00

Did you get the Error 6 problem solved on your Eppendorf 5424?

Report this Post
avatar placemarkReply by
wlin00

I also need a picture of the power resistor and how to fix it by soldering a wire to it. Thanks.

Report this Post

Page 1 of 3