Discussion Category:  Segmented Flow AnalyzersElemental AnalyzersChemistry Analyzers

CYanide by SKALAR

I have a SKALAR system that runs Free and Total CN.  At the moment the free is running fine but we can not get the total to run, does anyone have any ideas?

avatar placemark

Asked by

shy
Report this Post

Replies

EZkemReply by
EZkem

Hello

Can you tell me the problems you are experiencing?

Ken

Ken Earle,
EZkem
541-387-0343
www.EZkem.com
Report this Post
avatar placemarkReply by
Grandbend

Are you using the original lamp and glass coil for the uv digestor?

The proper lamp and correct coil material is important.

Report this Post
avatar placemarkReply by
shy

Just wanted to let you all know that we found the problem and our SKALAR is up and running. Thank God SKALAR has amazing customer service and they were able to trouble shoot the problem with us over the phone in a matter of minutes. It ended up being a clog in the lines comeing from the detector. It was just a small clog and it apeared like the flow was fine but it wasn't.

Report this Post
EZkemReply by
EZkem

I am glad they were able to help with such a simple problem.  Usually the problems with their Total Digestion System we solve on the Skalar systems by replacing the poorly designed glass digestor coils with teflon which has much better light penetration properties.

Ken Earle,
EZkem
541-387-0343
www.EZkem.com
Report this Post

Page 1 of 1