error 265

I have a problem with KF titrator. I can not fill beaker with Solvent manager. Display error 265. How can solve problem? Draining is OK, but fill with methanol not.

avatar placemark

Asked by

Edit
Report this Post
jburkman11Reply by
jburkman11

Hello,

Here is an answer from our Technical Support team:

Error 265 means that there is an i-chip error that requires service.  But this is unrelated to the solvent manager function.  If the solvent manager works to drain but not fill, it is very close to functioning.  I recommend using teflon tapes on the solvent and waste bottle threads to tighten everything better.  Everything must be tight, including in the titration beaker, the small needle hole stopper must be present.

Hope that helps!

Jennifer


Jennifer Burkman
METTLER TOLEDO
www.mt.com  