Discussion Category:  PCR / Thermal Cyclers

bio-rad cfx384

When the cfx384 is finished booting up, the message "storage card not found" shows up on the screen and the computer does not recognize the cfx384.

What to do?

avatar placemark

Asked by

mthalken
Report this Post

2 Replies

avatar placemarkReply by
Notsch

Quote mthalken

When the cfx384 is finished booting up, the message "storage card not found" shows up on the screen and the computer does not recognize the cfx384.

What to do?

I have the same issue but with the Bio Rad c1000 Touch Thermal Cycler. After rebooting it shows "storage card not found"

Can anybody help?

Report this Post
avatar placemarkReply by
Bodasafa

Hi,

I believe you both ahve the same problem.

The micro sd card has most likely lost its format. It has information on it required for the machine to boot up properly. The card is located on the back of the main display panel inside. You can take it out and read it on a PC. You will probably find its blank.