Discussion Category:  LC-Tandem MS

MS Scan Troubleshooting

Hello,I face a problem with the Quattro Premier XE Mass spectrometer in the "MS Scan" mode. The scan starts, the noise signal appears on the screen. Then a delay of 3 minutes occurs, afterwhich a straight line is drawn and the noise signal appears again on the screen.What may be the problem? Can anyone help? Thank you.

avatar placemark

Asked by

Lara
Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie

Lara,

Check your timed events table for the MS.  It sounds like you have a time segment set to a solvent delay or the MS turned off for those 3 minutes.

Dan


Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
Lara

I'll check this, thank you Dan! I think it might be the solvent delay because it does it repeatedly, even during the same run.

Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie

Lara,

Still would not ne a bad idea to check your timed events table, however I did not realize that this is recurring during a single run.  There might be something more to this.

Dan


Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post