Unable to calibrate

Hi,

I have only recently started using this machine (Seven multi). I have calibrated with no problem for the last 2 weeks. However with the last 2 days, I have been unable to calibrate the left probe. When I put the left probe into buffer 7 and press Cal, I get 3 broken lines and no value. i have checked the probe and it seems in good condition, please note this is a relatively new probe. Can anyone please troubleshoot for me. Thanks Katie

avatar placemark

Asked by

KCondon
Report this Post
LabWrenchEditorReply by
LabWrenchEditor

Please specify which model of the SevenMulti you are referring to.

Best Regards,
Kylie


Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com


Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
 
Report this Post
avatar placemarkReply by
KCondon

Hi Kylie,

The model is SevenMulti Neutral Meter S40.

Thanks,

Katie

Report this Post
avatar placemarkReply by
Sarah T

Do you get a reading when the electrode is placed in the buffer normally i.e. not trying to calibrate?

Sarah


Sarah Talbot BSc MSc MRSC

Method development and laboratory equipment sourcing specialist  
Report this Post
avatar placemarkReply by
KCondon

Hi Sarah,

I get no readings with buffer 7 and 4. I get 3 lines.

Thanks,

Katie

Report this Post
jburkman11Reply by
jburkman11

Hi Katie,

 

I spoke with our Technical Support Team and Michael answers:  

What is the model and serial number of the pH electrode? Perform the mV measurements below.

 

Using pH buffers 4, 7 and/or 10, make a note of the readings and compare the values to the following:

Buffer   mV .

4              180 ±20

7              0.0 ±20

10           -180 ±20

I hope this helps a little and I'm happy to ask Michael more if you have the model and serial number of the pH elecrode.  His email address is also uslabtechsupport@mt.com.  Thank you, Jennifer

 

 


Jennifer Burkman
METTLER TOLEDO
www.mt.com  
Report this Post
avatar placemarkReply by
KCondon

Hi Jennifer,

 

The model number of the probe is Meetler Toledo Inlab Epert Pro-ISM and the number on the cable is 3440397. I have read the following,

Buffer mV .

4     180 ±20  My reading was 2000mV and then cut out to 3 broken lines

7     0.0 ±20  My reading was -840.6mV

 

Thanks,

Katie

 

 

Report this Post
jburkman11Reply by
jburkman11

Hi Katie,

 

Michael replied and said that it is a dead electrode from 2013 and  unfortunately, you will need to replace.

 

Thank you,

 

Jennifer


Jennifer Burkman
METTLER TOLEDO
www.mt.com  