Discussion Category: Centrifuge
EMERGENCY OPEN
How do you do an emergency open on the spincron
Asked by
ACCESSING THE ROTOR IN CASE OF POWER FAILURE If the facility power fails only momentarily, the centrifuge will resume operation when power is restored and the rotor will return to set speed. However, if the rotor comes to a complete stop you will have to restart the run when the power is restored. In the event of an extended power failure, it may be necessary to trip the door-locking mechanism manually to remove the rotor and retrieve your sample. Action 1. Turn the power switch to off (O) and disconnect the power cord from the main power source. 2. Make sure that the rotor is not spinning. 3. Move the manual lock lever to the left (unlocked). Result No displays or LEDs are lit. No sound or vibration comes from the centrifuge. ! WARNING Never attempt to override the door interlock system while the rotor is spinning.
ActionResult4.Insert the supplied hexwrench (929317) into theoverride key slot in the leftside panel.5.Turn the hex wrench (either direction).The latch “clicks” or the door pops open.If the rotor is still spinning, close the door and wait until it stops before attempting to remove it.
"ACCESSING THE ROTOR IN CASE OF POWER FAILURE If the facility power fails only momentarily, the centrifuge will resume operation when power is restored and the rotor will return to set speed . However, if the rotor comes to a complete stop you will have to restart the run when the power is restored. In the event of an extended power failure, it may be necessary to trip the door-locking mechanism manually to remove the rotor and retrieve your sample. Action 1. Turn the power switch to off (O) and disconnect the power cord from the main power source. 2. Make sure that the rotor is not spinning. 3. Move the manual lock lever to the left (unlocked). Result No displays or LEDs are lit. No sound or vibration comes from the centrifuge. ! WARNING Never attempt to override the door interlock system while the rotor is spinning.
ActionResult4.Insert the supplied hexwrench (929317) into theoverride key slot in the leftside panel.5.Turn the hex wrench (either direction).The latch “clicks” or the door pops open.If the rotor is still spinning, close the door and wait until it stops before attempting to remove it."
I got it. Thanks for your help