Yet another mouse issue

We recently acquired a used DU-640, and I've been tasked with getting it functional.  It boots up and passes all its initial diagnostics, and at that point, the mouse pointer is in the top left corner of the screen.  I'm able to move the pointer all around the screen with the keyboard keys, but I can't click on the option to get beyond the results screen.  Is there a unique keyboard shortcut that can replicate clicking the mouse?  Neither of the enter keys, the function keys, or control/alt keys seemed to do anything.

I've been reading through the threads and I'm confident in my ability to modify the mouse's serial port with the jumper between pins.  What I'm wondering is if there is a way to tell if I have a motherboard revision that includes the fix.  Thanks.

Mulder

Asked by

Mulder
Report this Post

41 Replies

Chad S.Reply by
Chad S.
Do you have a printer hooked up? One thing to try if you do is make sure you have the printer on or disconnect the printer cable into the Beckman. Power cycle the system and see if that clears it up. It could be a simple bi-directional issue? This fixed the same problem I had once.   Chad  