"Lid" Error Message

Hey all,

I just got my machine running a few minutes ago after a minor circuit breaker repair. For some reason I keep getting the LID error after running the machine at 3200RPM for a couple minutes. The lid is well secured as far as I can tell. 

Any ideas what the reason might be for this message?

avatar placemark

Asked by

mrwright
Report this Post

41 Replies

avatar placemarkReply by
medphysteo

The lid may be mechanically latched but you might have an issue with the switch(es?) in the latch assembly, or possibly a loose connection to the circuit board (since you had it fully dismantled, it might be good to double-check). At the top of this page you'll see "View Thermo Scientific IEC Centra GP8 product details" under your post's title. Click on that link, then on the Documents tab where you'll find a service manual. 