Ms problem

Data subscriptions failed.

Are inbound ports blocked?

Is there right type of GC at the IP address? " how can i resolve this problem

avatar placemark

Asked by

Emy
Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie

Emy,

What models of GC and MS are you using, what Operating System and what version of ChemStation are you using?

 

Dan


Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
Emy

Gc 7890 ,5975ms

Chemestation b.01.06

Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie

Emy,

It sounds like you have lost communication with the GC, I would check all of the lan cables as well as the LAN hub/switch to ensure they are firmly connected.  After doing that I would reboot the GC and computer system, power up the computer first and then restart the GC.

If you are still having communication problems we would need to next verify communications with both the GC and MS from the computer

 

Dan

 

 


Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
ile

Hi, we have the same problem and we have no solution

Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie

What model of GC and MS are you using as well as ChemStation version and OS?

Did you check all of the lan cables as well as the LAN hub/switch to ensure they are firmly connected? As well as rebooting the system, powering up the computer first and then restart the GC.

If you are still having communication problems we would need to next verify communications with both the GC and MS from the computer

Dan