You do not mention the upper and lower pressure limits to which you have set the pump. As Bill has mentioned, you may have set the upper pressure limit at too low a value.
Please be aware that pump over pressure, may be also due to:-
-
The analytical or guard column being too ‘dirty’ or clogged with debris.
-
A blockage in the LC system tubing/mobile phase. Is your mobile phase filer clean, should your mobile phase be filtered.
-
Too high a flow rate for the analytical and guard columns.
-
Inadequate degassing of the mobile phase.
Please check out the above points.