Discussion Category:  High Pressure Gradient

Cecil adept ce4100 pump

The pump serial no 150-311, is indicating error e53 - pump over pressure! The instrument would not purged, and when the purge button is depressed, it gives the error e53, and shuts down the pump. The pump would not run even when the run button is pressed. What could cause this, and how do I overcome this problem problem?

J.o.a. Ogundeyi

Chief technologist

Mcrl university of ibadan

 

Asked by

20110609155000940

avatar placemark

Asked by

20110609155000940
Report this Post
LabWrenchEditorReply by
LabWrenchEditor

Over pressure alarm has been tripped. It is probably plugged or the alarm has been set too low.

Thank you to Bill Letter for providing this response to LabWrench


Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com


Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
 
Report this Post
LabWrenchEditorReply by
LabWrenchEditor

You may have to replace the alarm trigger (transducer).

Thank you to Bruce Neagle for providing this response to LabWrench


Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com


Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
 
Report this Post
avatar placemarkReply by
20110609155000940

Thank you very much,Bill Letter & Bruce Niggle. The pressure alarm was set too low, and it has been reset. The system in now ok.

Ogundeyi

Report this Post
Cecil InstrumentsReply by
Cecil Instruments

Dear Mr Ogundeyi,

Thank you for your recent enquiry.

You do not mention the upper and lower pressure limits to which you have set the pump.   As Bill has mentioned, you may have set the upper pressure limit at too low a value.

Please be aware that pump over pressure, may be also due to:-

  • The analytical or guard column being too ‘dirty’ or clogged with debris.
  • A blockage in the LC system tubing/mobile phase.  Is your mobile phase filer clean, should your mobile phase be filtered.
  • Too high a flow rate for the analytical and guard columns.
  • Inadequate degassing of the mobile phase. 

Please check out the above points.

If you have any further queries, please contact me directly, and provide your full contact details.

 

Kind regards,

Ade Kujore

Marketing

Cecil Instruments Limited

Milton Technical Centre, Cambridge CB24 6AZ United Kingdom

email:- ade.kujore@cecilinstruments.com

telephone:- +44 (0) 1223 420821

fax.:- +44 (0) 1223 420475

web site:- www.cecilinstruments.com

 

Registered Office: Unex House, Bourges Boulevard, Peterborough PE1 1NG,

United Kingdom  Registered Number 909536

 

Cecil AutoQuest LC autosamplers, for elegantly effective sampling.

http://www.cecilinstruments.com/adept-autosamplers.html

  


Email: ade.kujore@cecilinstruments.com

Telephone: +44 (0) 1223 420821 Fax.:- +44 (0) 1223 420475

Web site: www.cecilinstruments.com

Registered Office: Unex House, Bourges Boulevard, Peterborough PE1 1NG, United Kingdom Registered Number 909536  
Report this Post
avatar placemarkReply by
20110609155000940

Thank you ALL!

The upper pressure limit has been reset to 30mpa, while the lower pressure has been reset to 0mpa. The last user must have adjusted the settings, making the pump to trip!

Ogundeyi