DIFFICULTY IN THE MASS FILTER ELECTRONIC
MY 5973 MSD SHOWS DIFFICULTY IN THE MASS FILTER ELECTRONIC.
PLEASE CAN YOU HELP US RESOLVE THIS ISSUE.
HP/Agilent MSDs with "Difficulty in the Mass Filter Electronics" usually have a problem with either the Top Board, The Quadrupole Contacts, or the Quadrupoles (themselves). Rarely the Main board is the issue.
Use the proper procedure (read the manual) and cool and vent the instrument. Open the chamber and make sure all the connnections are made and no wires are touching each other or are grounded. Thermocouples and RTDs will be grounded, of course.
Pump the MSD back down and give it another try. If this does not work, seek the help of a qualified service agent .... you probably have a problem.
Scott Niemann
CSS Analytical Co Inc.
www.cssco.com
800-277-5455
Reply byThere are several things that could be causing this problem the most basic of which is lack of good vacuum is this a turbo model or difusion pump model? Do you have a ion gauge controller on it if so what is the vacuum that you're reading? 9 times outta 10 lack of vacuum causes this error.
As sharpie said sometimes this may be an electronics issue, but typically it is not and there are further troubleshooting steps that should be taken after ensuring your vacuum is good before you jump to the conclusion that it is an electronics issue.
Dan
Dan Ste.Marie
RJ2 Instrument Services
Erie, PA
www.rj2.biz
Thank you Scott, Sharpie and Dan. I have the same error message from a 5973A with a diffusion pump. Although there's no ion Guage controller in the unit, the vacuum appears to be good. I have checked that all contacts and connections on the quad are okay. when I click on the Optimize the Quad frequency in the Execute menu, rather than get the dialog box where I would enter a mass value so as to get a readback, what I get is the result of the MS I/O test. Could this be a SideBoard problem? I will appreciate your insight. Thanks.
Muyi
AAS Ltd
Reply byMuyi,
How are you determining that the vacuum looks good? What is the foreline pressure and what levels is the diff pump oil and is it clear? Is your rough pump oil at an adequate level and is the rough pump gurgling excessively?
Dan
Dan Ste.Marie
RJ2 Instrument Services
Erie, PA
www.rj2.biz
Thank you, Dan. The oil in the diffusion pump is clear and just a little above the cold level when the instrument is shut down. The oil in the rough pump is about three-quarters filled up. The vacuum chamber seals tight when the instrument is on and the rough pump does not give any execessive gurgling sound. It only gives the usual gurgling sound when the instrument is just started up after venting/shutdown for sometime, and with the vacuum chamber pressed tight by hand, the gurgling stops after about 1-2 minutes. These are the reasons I think the vacuum may be good even though there is no vacuum gauge to confirm this. I am unable to check the vacuum status from the software. Once I click on Vacuum status in the software, I get the "difficulty in mass filter electronics" message. I am also unable to optimize the quad frequency - the readbacks from this would have helped one to have an idea about the sideboard.. When I click on the Optimize Quad frequency, i only get the results of the MS I/O test. Thinking this could be a software corruption issue, I reinsalled the software but this did not help. Unforunately, I do not have a spare sideboard to test with. I just want to be sure before I commit to purchasing a new sideboard
Reply byMuyi,
It sounds like you do have a more serious issue. Based on what you stated I would have to agree that the vacuum is probably good ( or good enough to be able to get some results). I would talk to the service department at one of the reputable instrument vendors as they may work with you on the purchase of a sideboard as there is a possibility it could be the main board.
Dan
Dan Ste.Marie
RJ2 Instrument Services
Erie, PA
www.rj2.biz
Thanks, Dan. I got a 5975MS sideboard and I just put it in anyway just to check. It seemed to work because I didn't get the error message again. I could see my vacuum status and foreline pressure was about 46mTorr. But the MS wouldn't complete autotune; said "Peaks are too narrow". It wouldn't even complete air/water check, it would stop with the message that "Peak widths are inconsistent". I have reloaded a fresh tune file from the config menu; plugged the transfer line; checked for leaks, replaced the EM. Nothing seemed to work. I suspect this may be because am using a 5975 sideboard. I couldn't completely optimize the quad frequency- I got 66mv for the 100 amu but could not get below 470mv for the 650amu. So I think I simply have to get the right sideboard for the 5973MS.
Reply byMuyi,
Try retuning the quads but instead of trying to lower the reading at m/z as much as possible leave it around ~80 and then only CHECK the reading when set to 650 m/z. It should be GREATER then 500. Then try retuning. I have never run a system with a 75 sideboard and 73 mainboard. I believe that the 73 mainboard will have to be one of the later fast electronics board. Did you get a log amp/sideboard mismatch error?
Dan
Dan Ste.Marie
RJ2 Instrument Services
Erie, PA
www.rj2.biz
Don't forget the 5973 side board and log amp (small board on the main board) much match.... i.e. be compatible.
Scott Niemann
CSS Analytical Co Inc.
www.cssco.com
800-277-5455
Thanks, Dan....and Scott. No, I didn't get the log amp mismatch message, so I suppose the 73 is the fast electronics version. Dan, I will proceed with the Quad as you have suggested. Many thanks.