slope for 3 point pH calibration
slope value that we had for the 3-point pH calibration is 91.3%, for the seven multi S47K how the value range for the slope and how to adjustment?
Discussion Category: pH Meters
slope value that we had for the 3-point pH calibration is 91.3%, for the seven multi S47K how the value range for the slope and how to adjustment?
Asked by
Hello,
Nancy, from our Tech Support team, answers below:
The slope is calculated from the actual mV values that are the raw signal for the pH meter. The mV are converted to pH (a concentration). The only way to change the mV readings, and therefore the slope, is to either buy a new electrode, or try to maintain the electrode that we have. A slope from 90% to 95% is actually good for "routine" measurements, and a slope of 95 to 105% is good for "accurate" measurements.