Discussion Category:  Centrifuge

Error drive box

At the moment we have 10 Sorvall RC12BP machines and now have 3 machines not operating because it indicates error drive box.

You can send us service manual?

avatar placemark

Asked by

tonthatdung
Report this Post
LabWrenchEditorReply by
LabWrenchEditor

Hi tonthatdung,

Attached is the user's manual for the Thermo Scientific Sorvall® RC12BP™. The bottom of page 3-19 may help (drive box fault).

Sorvall Manual.pdf

Thank You,
Kylie


Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com


Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
 
Report this Post
avatar placemarkReply by
tonthatdung

Hi Kylie,

In the Sorvall manual has box drive fault but hasn't guide to repair it. Can you send for me Diagram about Box drive ?

Thank you and best regard.

Report this Post
avatar placemarkReply by
Emma_W

Hi tonthatdung,

Your local Thermo Scientific team will be able to help you out. Just go to the Contact Us section of our site, and select "service", then your country and the appropriate category - this will give you your local office's contact details.

https://www.thermoscientific.com/ecomm/servlet/leadentryform?storeId=11152&langId=-1&from=mcu

Report this Post