Negative values on chromatogram for Peaks

When running Dionex ICS-2000, I am getting negative values for my samples. What causes this?

avatar placemark

Asked by

Samuel
Report this Post
LabWrenchEditorReply by
LabWrenchEditor

Negative values for samples is caused by having a calibration curve with offset.  The nature of the curve near the origin if it goes negative at low values will give negative result values.

Thank You to Dionex for providing this response to LabWrench


Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com


Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
 
Report this Post
cloudyarunReply by
cloudyarun

At what time negative peak is coming. First peak will be the water dip peak

 

Be happy  
Report this Post