Discussion Category:  PCR / Thermal Cyclers

Veriti 96

We receive the following error, wondering if anyone knows a fix without sending it out?

Error Number: 0x8013 Data: 0x00,0x00,0x00,0x00

This error is unrecoverable. Please contact the technical support and restart your instrument.

avatar placemark

Asked by

crsimons
Report this Post
avatar placemarkReply by
tom1991
Hi, I had the same issue x8013 and I measured the resistance by @Bodasafa, the resistance are in the right range, but the system still report the X8013 issue and can not operate. It is not the heat cover's wire issue. I think most of the 0x8013 issue is related to the heat control board, not the wiring. If the system can power on, it means the fuse is good. You need to check the control board and fix it.