How to Sanitize a Millipore AFS 3D system

I have a AFS 3d water system that feeds pur water to a chemistry analyzer, does any one know how to sanitize this system??

avatar placemark

Asked by

BL
Report this Post

1 Reply

avatar placemarkReply by
Paul Moyer

The system needs to be sanitized using chlorine dioxide, do not use chlorine bleach or any other type of cleaner as they might cause bad results.

The chlorine dioxide is very specific to the millipore system, I would suggest calling Millipore before attempting to disinfect  the tank and the internal parts of the system. This service will require some tubing connections to be removed and closed off as chloring dioxde kills bacteria as a gas.

I have done numerous systems and do not recommend doing so without training.

Paul M.