Discussion Category:  Cell Counters

Check sample level

Hi everyone, currently my machine pop-up the error " check sample level ". Already check the level of the sample ( ok ), Drain and refill the sistem but the same error still pop-up. Anyone can assist?

avatar placemark

Asked by

Jadd Juhari
Report this Post
avatar placemarkReply by
lbrow028

I am having this same issue.  Did you find out why the error and how to fix?

Report this Post
avatar placemarkReply by
labotronics.mary
The “Check sample level” error may be caused by a faulty sample-level sensor, a blocked probe, or air bubbles in the sampling line, even when the sample volume is sufficient. Since draining and refilling the system did not resolve the issue, check the probe and tubing according to the manufacturer’s instructions. If the error persists, contact technical support and provide the machine’s make and model for accurate troubleshooting.
Report this Post