Discussion Category:  GC-Single Quad/MSD

5975C Smart card issues

Wondering if anyone has this same problem.  We have a 5975C that will occasionally randomly cause one of the target ions we are looking for to "disappear".  For a target peak the secondary and tertiary ions are present and fine but the primary looks like a "carrot" and/or baselines several times during the peak elution window.  In the past I have replaced the smart card with the thoughts that the signal is dropping or corrupt.  This will solve the problem but it seems to reoccur after 6months to a year. 

avatar placemark

Asked by

nbhllwy
Report this Post
avatar placemarkReply by
sharpie

Your comment " during the peak elution window" makes me wonder if the problem  you experience presents during/in the "real time" window or is actually present in the data file ?


Scott Niemann

CSS Analytical Co Inc.

www.cssco.com 

800-277-5455  
Report this Post
avatar placemarkReply by
nbhllwy

It's in the real data file. randomly, different ion every time.

Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie
If you are not having any other communications issues, I would suggest that the problem is more likely with the signal cable (black coax cable) or its connections. I would verify that both ends are clean and firmly connected. I have had to replace these in the past. Dan

Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
ECS

Are you in scan/sim mode?  

Report this Post
avatar placemarkReply by
nbhllwy

Scan mode

Report this Post