Discussion Category: GC Systems
Shimadzu GC-2010 configuration lost when changing battery
The GC displayed an error of battery low. Then I have stopped the equipment and restart it again, and the language is now in japanese. We have changed the battery now but need to restore the equipment to the normal operation settings first of all return the equipment to english language. Please help me with this.
Thank you,
Adolfo
Reply byAdolfo,
It should be FUNC then Item 6, then Item 9 and then Item 1 under the FUNC submenus.
Dan
Dan Ste.Marie
RJ2 Instrument Services
Erie, PA
www.rj2.biz
Greethings from Mexico City.
My name is Idelfonso Hernández Cipres.
Please, send me more information, my Email is aysa96@hotmail.com
It is necesary to restore the communication.
The communication beetween pc and GC 2010 is lost.
It is necesary a program and communication cable RS 232, beetween pc and GC 2010.
IHC
Hello Dan,
Thank you for your answer. The problem is that the equipment is now displaying only in japanese and I am not able to see the menus in english. I will send you the photos of this problem, but when I am in the lab.
Thank you,
Adolfo
Reply byDan Ste.Marie
RJ2 Instrument Services
Erie, PA
www.rj2.biz
Hello Dan,
Until today I was able to go to the lab. I have used the keystrokes you gave to me, and now I have the equipment again in English. Now I am configuring again the GC, I hope that every thing goes OK, but in case I have trouble, I will cask to you again.
Thank you a lot it saves us in the lab a big bunch on dollars. Have a nice day.
Regards,
Adolfo
Reply byDan Ste.Marie
RJ2 Instrument Services
Erie, PA
www.rj2.biz
Hello Dan,
The configuration of the GC was lost, so I am working to reconfigure it. My GC has three detectors: FID, ECD and FPD and two Split/splitless injectors. But now when I check the INJ button for injectors the display shows : Inj port is not installed and for detectors the same, when I press DET it says Detector is not installed.
I went then FUNC pressed 7 Service/Maintenance and then went to 1 Installation (Position)
Then I have configured the position of Injectors and detectors:
Port Type Position
INJ1 SPL 6
INJ2 SPL 7
DET1 FID 1
DET2 ECD 2
AUX3 FPD 4
AUX4 ------- ---------
AUX5 -------- ---------
I have selected the position according to a table in the GC manual.
I have checked in the manual of the GC but I cannot find the proper explanation.
I don't know how to configure this detectors and injectors., can you help me with this?
Thank you,
Adolfo
Reply byAdolfo,
Have you configured the analyical lines on the system. The detectors and inlets will not be active until they are installed and configured as part of an analytical line.
Dan
Dan Ste.Marie
RJ2 Instrument Services
Erie, PA
www.rj2.biz
Hello Dan, I don´t know what is an Analytical line. I have an operation manual but I don't find or I don't understand what is an analytical line. Can you help me with this?
Adolfo