Discussion Category:  GC Systems

Shimadzu GC-2010 configuration lost when changing battery

The GC displayed an error of battery low. Then I have stopped the equipment and restart it again, and the language is now in japanese. We have changed the battery now but need to restore the equipment to the normal operation settings first of all return the equipment to english language. Please help me with this.

Thank you,

Adolfo

avatar placemark

Asked by

Adolfo Alvarez
Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie

Adolfo,

 

It should be FUNC then Item 6, then Item 9 and then Item 1 under the FUNC submenus.

 

Dan


Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
I. Hernández Cipres

Greethings from Mexico City.

My name is Idelfonso Hernández Cipres.

Please, send me more information, my Email is     aysa96@hotmail.com

It is necesary to restore the communication.

The communication beetween pc and GC 2010 is lost.

It is necesary a program and communication cable RS 232, beetween pc and GC 2010.

 

      IHC

Report this Post
avatar placemarkReply by
Adolfo Alvarez

Hello Dan,

Thank you for your answer. The problem is that the equipment is now displaying only in japanese and I am not able to see the menus in english. I will send you the photos of this problem, but when I am in the lab.

Thank you,

Adolfo

Report this Post
avatar placemarkReply by
Andino
Dear Adolfo,   Maybe it's not the best and technic answer, but in this case, I will use the Google Translator App from a cell phone, it can translate letters from pictures.   Take a photo of the Japanese screen and configure the translator to read the screen in Japanse and translate it to English, in that way you can naviagte until the Language Configuration Screen, and return the GC to English.   Instructions to translate images. https://support.google.com/translate/answer/6142483?hl=en   Successes.   Andrés
Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie
Adolfo, Even though it's in Japanese you can still enter that sequence of commands to change it to English. Dan

Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
Adolfo Alvarez

Hello Dan,

Until today I was able to go to the lab. I have used the keystrokes you gave to me, and now I have the equipment again in English. Now I am configuring again the GC, I hope that every thing goes OK, but in case I have trouble, I will cask to you again.

Thank you a lot it saves us in the lab a big bunch on dollars. Have a nice day.

Regards,

Adolfo

Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie
Adolfo, I'm glad I could help out. Dan

Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
Adolfo Alvarez

Hello Dan,

The configuration of the GC was lost, so I am working to reconfigure it. My GC has three detectors: FID, ECD and FPD and two Split/splitless injectors. But now when I check the INJ button for injectors the display shows : Inj port is not installed and for detectors the same, when I press DET it says Detector is not installed. 

I went then FUNC pressed 7 Service/Maintenance and then went to 1 Installation (Position)

Then I have configured the position of Injectors and detectors:

Port               Type               Position

INJ1                 SPL                   6

INJ2                 SPL                   7

DET1               FID                    1

DET2               ECD                 2

AUX3              FPD                  4

AUX4             -------              ---------

AUX5             --------            ---------

I have selected the position according to a table in the GC manual.

I have checked in the manual of the GC but I cannot find the proper explanation.

I don't know how to configure this detectors and injectors., can you help me with this?

Thank you,

Adolfo

Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie

Adolfo,

Have you configured the analyical lines on the system.  The detectors and inlets will not be active until they are installed and configured as part of an analytical line.

Dan


Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
Adolfo Alvarez

Hello Dan, I don´t know what is an Analytical line. I have an operation manual but I don't find or I don't understand what is an analytical line. Can you help me with this?

Adolfo

Report this Post