DANI master TOF GCMS

Hi! I have a problem in my  DANI master TOF GCMS, he fail autotune check tests and leak check tests. It is a new, multiplayer value recomented is 2500-2600 for a start, but he fail check with 2800. leak check for 5 tests is 2 pass and 3 fail. please help. Any one have problems with new equpment?  maybe it is a used mashine?

avatar placemark

Asked by

georgiy23
Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie
As long as you have leaks you will not be able to successfully tune the instrument. Concentrate on finding and resolving the leaks first. When you have a leak and attempt to tune, the algorithms will boost the EMV to compensate for the drop in sensitivity due to the leak. Dan

Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
georgiy23

we check the leaks with agilent service manager. in column, liner it is ok no leaks but any time leack check is ok

 

Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie
it is unclear from your reply whether or not you resolve the leaks. Were you able to resolve the leaks?