Both light bulbs won't work
Power light is on, neither bulb will work. Is there a fuse I can check and where is it? Thanks

Are you using the integrated power supply or the external unit?
Thank you to Trevor B for providing this response to LabWrench via LinkedIn
Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com
Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn

I would contact your local Zeiss organization and they will be able to help.
Thank you to Simon K for providing this response to LabWrench via LinkedIn
Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com
Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn

Please write to our new central support email microscopy@zeiss.com with all necessary information and we'll get in contact with you immediately.
Thank you to Johannes A (Online Marketing and Communications, Carl Zeiss MicroImaging) for providing this response to LabWrench via LinkedIn
Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com
Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn

Is very difficult that both bulbs are faulty.
May be the stemi have problems in the keypad. +/- buttons do not work. Press the central button. I believe M. With this you can change the illumination.
Good luck.
Thank you to Juan A for providing this response to LabWrench via Linked
Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com
Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn

First, replace the two bulbs with two known good bulbs.
If the above does not work, then replace the internal PC board controller and the external power supply.
000000-1054-897 electronic control board $80.30
000000-0522-133 power supply 12V/2,5A $47.40
and
000000-0523-824 adapter cable $23.40
Thank you to Edward M (National Sales Manager E& R Microscopy at Carl Zeiss, Inc.)for providing this response to LabWrench via LinkedIn