Check connections: touch screen, CPU board and fermentation card.
Try reloading the software.
Once this process is initiated, the screen may be blank and appear to be doing nothing. Do not interrupt or restart the system! This process will take about 10 minutes to complete.
· Download update software from Eppendorf web site.
· Copy to USB drive.
· Connect USB drive to connector on rear panel of control cabinet.
· Turn on mains power.
· After several minutes, the screen calibration screen will appear. Follow instructions to calibrate the touch screen.
· Software loader screen will appear.
· Select the software that you wish to load. If no software shown press refresh to bring up software. Select software required. Press load update.
· When update complete. Go to setup tab and select hardware setup.
· Setup hardware as described in the manual.