reservoir sensor level

Hi,

our reservoir sensor level from our Milli-q Integral 5 is taking wrong readings, as 140% when is actually almost empty.. so, how can I reset or calibrate this kind of sensor?

Thanks for any help!

Isaias.

avatar placemark

Asked by

Isaias
Report this Post
avatar placemarkReply by
Paul Moyer

If you have the small beige tank attached, the float on the switch will swell and cause it to stick to the top. It would be best to replace the sensor at this point.

Report this Post
avatar placemarkReply by
Martha Sturgis

Hello,

I guess the problem is due to some technical error, you should visit Milli-Q® Integral 5 Water Purification System official page where you would find numerous documents and links to help you with this issue. You could also contact their technical support on 1800-209-9939.

Hope that helps. Thanks.

Report this Post
avatar placemarkReply by
MilliporeSigma Support

Open the top of the tank.  Remove the level sensor and float.  The float either fell off the bottom of the level sensor, or was so close to the bottom of the sensor that it was not reading correctly. 

 

Put the level sensor on the bench top, move the float half way up, what does it read for tank level?  50%? 

 

Call us at (888) 645-5478 for more help.