How do you disassemble the front of the instrument to take apart to lubricate the elevator assembly.
Discussion Category: Particle Size Analyzers
How do you disassemble the front of the instrument to take apart to lubricate the elevator assembly.
Asked by
Hi Ruthie,
Please contact Jason Noga at jason.noga@microtrac.com. He may be able to assist you with the issue you are having.
Best Regards,
Kylie
This is a pretty easy process. You just have to take out the 8 screws along the edge of the front panel. Carefully pull out the elevator assembly. There will be wires connected to it, so just ease it out enough to lubricate the screw. Use a Teflon grease, not an oil to lubricate. Push the elevator back in and tighten up all the screws and then run the elevator up and down a few times.
Thank you to Joe from Micromeritics for providing this response to LabWrench