Thermo Nicolet Avatar 360 FTIR

The above FTIR has not been used for over three years. Everything seems to be working and all diagnotic test show the green tick, however the performance test just shows bad scans. I have checked inside the compartment and the laser and source are working, although checking the laser light is not visible from the sample compartment. I do not wish to spend a huge amount to get it running, any help I would be very greatful.

Regards Allan

avatar placemark

Asked by

noddy46
Report this Post
avatar placemarkReply by
Leo_tech

Hi Allan,

have you tried another scan speed?

Report this Post
avatar placemarkReply by
noddy46

Thanks very much for your reply, yes I have tried various scan speeds. Since my last post I have found that the light is visible from the sample compartment but when using the polystyrene standard for example, the light seems to be too high, not going through the round window. I have reset and aligned the instument but its still too high.

Regards Allan

Report this Post
avatar placemarkReply by
Leo_tech

Did you run an "auto align" procedure?

Can you add here an interferogram and diagnostics screenshot?

Regards

Leo

Report this Post
avatar placemarkReply by
noddy46

Hi Leo,

Under the diagnostics everthing is between the normal level (green tick) however the advance diagnostics the interferogram shows too many bad scans, which seems to point to the detector, hope this helps.

Regards Allan

Report this Post