Discussion Category:  GC Systems

vent the effluents from the detector

How should I insert the vent tube of the FID detector, what material should I use? .for the dangers of working with pesticides

 

Rebozzz

Asked by

Rebozzz
Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie
The FID is a destructive detector so not much really is left into the vent gas from the detector. In addition you have a hydrogen flame so you really can't have anything directly connected. Your best option is to install a snorkel vent or extractor vent and position it above the detector. Dan

Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
RebozzzReply by
Rebozzz

THANKS, but it is exactly I refer. My detector equipment needs something that makes the effluents flow out to the extractor hood, I already have a Teflon hose installed, but I need to know what material I would use to attach it to it, because the manual says so.


   Moises Bozeta

Chemist in Organics Lab  
Report this Post