Discussion Category:  Hot Plate Stirrers

error 12

I have a C-MAG HS7 that is displaying error 12. Would you happen to know what that error is? It is not in the manual.

avatar placemark

Asked by

Aaron2184
Report this Post
avatar placemarkReply by
Hadaro

Hi. I hope it helps you. I've replaced the switch found in the picture and it fixed the error 12

Ika error 12https://imgur.com/a/0ybcT

 

Ika Error 12

 

Report this Post
avatar placemarkReply by
Mayur
Hello I am getting error11 on IKA CMAG HS7 , when i unplug mains earth from pcb Instrument works normally (not recommended as it passes leakage current over body of Instrument) But reconneting Earth wire doesn't resolve error 11 Let me know if you find anything
Report this Post
avatar placemarkReply by
Fonko

i have to changed the switch, but the alarm still occured.

any tips ?

Report this Post
avatar placemarkReply by
yuhntnanto848565

Thermocouple is broken

Report this Post
avatar placemarkReply by
lagutin
everything is much simpler: replace the motor and the device will work . Glory to Jesus Christ !