Discussion Category:  Lab Pumps

Welch Vac pump Rebuild

Hello fellows techs. Has anyone rebuilt a Welsh Chemstar 1402N . I'm sure I put it together right, but now it won't turn by hand. I'm suspecting the E-Clip or Thrust Washer in front of intake rotor. Any ideas? Thanks

avatar placemark

Asked by

Leon1518
Report this Post
avatar placemarkReply by
bberg7794

Hi Leon,

I have rebuilt a Welch 1400, a 1400N Chemstar and also a 1376.  I encountered a similar problem to what you are describing with the 1376.  When I disassembled it, there was a thrust washer but no thrust discs installed.  Since I was "rebuilding" the pump, I installed a new thrust washer in place of the original one.  The original was thinner with visible machining type wear marks.  After I had everything back together and torqued, the pump would turn over by hand, but it was difficult and obviously not correct.  I removed the end plate and also the new thrust washer, then reinstalled the original thrust washer and torqued everything back together again-then it turned over smoothly.  That was about a year ago and this pump is still running happily.

Hope that helps,

Brian

Report this Post
avatar placemarkReply by
Leon1518

Thanks for replying and the information. I'll give that a try.

Report this Post