Discussion Category:  UV/VIS Detector

Cannot connect to CBM-20A

The spd-m20a is shown as an unregistered device in the cbm-20a web interface therefore I cannot configure it. Does anyone know how to register a device? I have also used the software analyst 1.6, but it doesn't even let me choose a pda detector.
avatar placemark

Asked by

Amingh92
Report this Post

4 Replies

avatar placemarkReply by
ABUBAKR

Hi, you can configure through a web prowser by entering the following IP address 192.168.200.99  and log in

Report this Post
avatar placemarkReply by
Amingh92

Hi Abubakr Do you know what exactly I need to do? The Shimadzu pda is shown as an unregistered system in 192.168.200.99 None of the options on the browser give me the choice to register the PDA

Report this Post
avatar placemarkReply by
ABUBAKR

Did you try 192.168.200.98 because pda canot be configured through cbm page it has its own page

Report this Post
avatar placemarkReply by
Amingh92

Yes I did, and it worked. Thank you

Turns out, all I had to do was to set PDA`s group name to the same as CBM.

Now I cannot get a chromatogram from the PDA in CBM analysis page. Do you have any idea how PDA can export the data in the form of a chromatogram?